My trident has been giving me very inconsistent readings for Alk. The combined tests came out good but for some reason the alk reading would shoot 0.2-0.3dkh up when performing the alk only tests. I opened a ticket with Neptune and they had me set the schedule for 12 tests a day and my graph looked like a saw. They guided me through these steps so, if you're having this issue, please get with them first because you don't want to perform these steps on your own without their guidance and end up breaking the thing. Turned out the cuvette was dirty. I had run my system without filter socks for some time and that maybe have contributed to it. Just to be sure, Im never taking my filters out again. Tech support told me how to open the unit and take the cuvette assembly out. I cleaned it with qtips and and now it's back to normal.
This is the view of the cuvette from the front. You can see a film on the front and sediment on the bottom.
This is the view of the cuvette from the top down. There was that same film on all four sides. The photosensors are on the side.
And here are the qtips
This is the view of the cuvette from the front. You can see a film on the front and sediment on the bottom.
This is the view of the cuvette from the top down. There was that same film on all four sides. The photosensors are on the side.
And here are the qtips