For years I've used refractometer calibration fluid to calibrate my Vitalsine SR-6 refractometer. It's a great unit and I've had it for 14+ years. I was looking for some information about the Automatic Temperature Compensation (ATC) feature and went looking for the instructions. Turns out, the manufacture of this unit suggests calibration with just plan water and adjusting it to 0. You can read the calibration instructions through the link:
https://pentairaes.com/vital-sine.html
What gives? I was curious so I did a little experiment. I calibrated my Hanna Salinity checker (to 35 ppt), tested the refractometer vs. the salinity checker and got some interesting results. When I calibrated the refractometer with RODI water, the reading was 35 ppt. When I calibrated the refractomer with the 35 ppt calibration fluid, the reading I got was closer to 37 ppt.
When I calibrate the refractomer with plain water, per the instructions, the reading lined up with the freshly calibrated Hanna Salinity checker. When I calibrate the refractomer with the 35 ppt calibration solution, the reading was higher than the freshly calibrated Hanna Salinity checker. Should I be calibrating the refractometer with the RODI water moving forward? Have my tank been running closer to 37 ppt all these years? What do you think?
https://pentairaes.com/vital-sine.html
What gives? I was curious so I did a little experiment. I calibrated my Hanna Salinity checker (to 35 ppt), tested the refractometer vs. the salinity checker and got some interesting results. When I calibrated the refractometer with RODI water, the reading was 35 ppt. When I calibrated the refractomer with the 35 ppt calibration fluid, the reading I got was closer to 37 ppt.
When I calibrate the refractomer with plain water, per the instructions, the reading lined up with the freshly calibrated Hanna Salinity checker. When I calibrate the refractomer with the 35 ppt calibration solution, the reading was higher than the freshly calibrated Hanna Salinity checker. Should I be calibrating the refractometer with the RODI water moving forward? Have my tank been running closer to 37 ppt all these years? What do you think?