Danh;1092955 wrote:
anit77;1090206 said:
Yes, it was 53mS. /QUOTE]
Why should I pay $10-15 for calibration fluid when I can just use water sold to me a 35g/l and confirmed with my refractometer. Can't I just calibrate with that water??
You can't use the Neptune stuff to calibrate a refractometer. It will always show lower on it. Also yes evaporated fluid will give you a bad result.
The Neptune calibration fluid is only $3, don't know where your finding it for that high a price.
Float the bag in your sump for 10-15 minutes before you open it. If you have a decent size measuring cup fill it about 3/4 with water from the sump an hang it from a weir in the sump. If you have a temp probe connected to the PM2 throw it in the measuring cup. Clean the Condux probe with rodi water and dry, carefully. Start the calibration in the Apex and set the temp comp to 2.2 (Only if you have a temp probe connected to the PM2). Cut the corner of the bag and insert the probe at an angle with the air port facing up and swish it around in there. Then put the bag and probe into the measuring cup,<u> making sure no water from the cup gets into the bag.</u> Now let it sit for 20 min to a half hour. When you come back to it again swish the probe in the bag and see if the numbers change. If it's still changes after a couple seconds let it sit some more. Once it doesn't change for more than 20 seconds complete the calibration in Apex.
Now put the probe in an area that is free of bubbles in your sump and see what the reading is. If the reading starts getting low again use your hand to swish some water at the probe from underneath. These things are very sensitive to air bubbles.