Problem solved!!!
Issue was that my refractomer was being stored and used at different temperatures. I would guess the total swing in temp for the tool was 10 degrees. Hottest at night before lights go off and coolest in the morning.
The difference caused a 0.005 ish change in the reading.
Edit: The warmer the tool the farther the reading dropped and the cooler the temp the higher the reading became.